Thanks for any replies in advance.
I have a list of RRFs created on one instrument.
For similar column and flow rates - would the RRFs be transferable from
one instrument to another?
Regards
WK
Advertisement
Discussions about GC and other "gas phase" separation techniques.
How similar, and how much discrepancy can you tolerate ? Are they different instruments from the same manufacturer, from different manufacturers ?, AND most important of all, way more important than columns and flow rates - do they have the same detector ?Hi All,
Thanks for any replies in advance.
I have a list of RRFs created on one instrument.
For similar column and flow rates - would the RRFs be transferable from
one instrument to another?
Regards
WK
So you have a three-point calibration that you want to transfer to another instrument. Bad plan I would say, and to run 3 points on the new instrument seems like way less stress than worrying if the two instruments behave the same.Sorry Peter :
I have calculated average RRFs from 3 levels (against an internal standard).
I have then calculated %s in test samples using the RRFs and adding the internal standard used in calculating the RRFs.
This works reasonably well to +/- 5% of actual % (e.g. 0.95 - 1.05% for 1.00%).
Which is fine for the application.
I was just curious how well they would transfer to different FIDs.
Regards
WK
Hi WKHi Peter/Don,
Thanks again for the replies.
I am following "Intelligent use of relative response factors in GCFID" in Chromatography Today May/June 2012.
I must stress this is not for legislative/qc work - it is only to replace %area as a guide to %w/w.
Only as an in-house guide for creating a start for a formulation that will be adjusted and modified within legislative boundaries.
e.g. component 1 is 1.00% area but in reality 0.05% w/w.
I take on board your comments on injector liners - I use my manually wool packed liners -
I have trouble with the focus type liners and only pack empty liners the way Perkin Elmer recommend.
I would love to buy off the shelf fully packed liners with wool in place !
Best Regards
WK
Hi WKHi Peter,
The solvent used mainly is ethanol (sometimes with water) and makes the differences seen between %w/w and area%.
Regards
WK
I have a dream…. in your situation ethanol may be used as Internal Standard.Hi Peter,
The solvent used mainly is ethanol (sometimes with water) and makes the differences seen between %w/w and area%.
Regards
WK