My 2 cents: I agree with one of the statements above that it is unlikely to be able to measure everything with one TCD, for the reasons stated -- whichever carrier you choose, you're going to have a hard time differentiating analytes with a similar mass to the carrier (i.e. helium/h2, argon/air). HID was suggested, but perhaps DID or PDID is more suitable.
Your other problem is separating everything without trashing your column -- The solution I have used in the past is actually columns in series, rather than parallel. You inject the sample onto a column (Porapak, PLOT or whatever). Non retained components (permanent gases) blast through and get to the second column, a molesieve. At this point you have a 4-port valve programmed that isolates the molesieve while the heavier stuff elutes (CO2, water, everything else). After everything is done eluting, you switch back to the molesieve and let the permanent gases elute.
As mentioned elsewhere, you don't want to load up a molesieve with CO2, moisture, etc. It will wreck the chromatography for everything else and require a lengthy bakeout (overnight is pretty standard).
VICI is a standard supplier of the 4-port valves along with actuators (either pneumatic or electronic). I've been buying packed columns from Ohio Valley, though there are many other good vendors I'm sure.
Good luck,