Agilent 7890A infinite incubation (Headspace)

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello, to start sorry for my bad English, I am French, but I will try to be most clearly possible.

Terms:
- 40°C, 5°C/min => 240°C (40mn)
- Column dB-Wax 30m*0.32mm*0.5µm
- H2 0.8mL/min, constant Pressure 29.09
- FID detector
- Time of incubation of the samples: 180mn

Software:
Management Frontier runner: Maestro
Aquisition management: Ezchrome

I have a problem with my GC, I work in Headspace and when I launch my sequence of 25 samples (26h ~ of run) my device incubates my sample N°6 ad infinitum (the 5 first did not have any problem in the injection and the aquisition).

Do you have an idea of why? I wanted to look at if there were a “Run Prep” but I read that it was automatic.

Thank you in advance
Sometimes, when a new sample has to be loaded into the headspace oven at the same time as a sample is being injected to the GC the whole system freezes. Try reducing or increasing the incubation time by two minutes to avoid this clash.

Two comments on your method; do you really need such a long incubation time ? and your carrier gas flow rate is much too slow - you will get better separations in less time if you use 2 ml/min.

Peter
Peter Apps
Ok I will try to start again the sequence by changing the time of incubation of one or two minutes. I would tell you if it goes better

For my method, I work on the analysis of the headspace of a wine of synthesis, my time of incubation was determined compared to the moment when I am with balance, it is my optimal time of incubation.

For my flow, I must be with a swiftness of 37.2cm/sec with constant flow, if it is higher I have besides Co-elutions in my spectrum.
The linear flow that you are aiming at is for helium. For hydrogen the optimum is 50 cm/sec. I think that you need to check your flow anyway - given your inlet pressure at your starting temperature your volume flow is about 1.6 ml/min, which gives a linear velocity of 31.4 cm/s. If your volume flow really is 0.8 ml/min your flow rates are appropriate for nitrogen carrier, which is very slow.

Peter
Peter Apps
4 posts Page 1 of 1

Who is online

In total there are 63 users online :: 0 registered, 0 hidden and 63 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 63 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry