I am attempting to clean an HPLC column, and the protocol recommends washing with 1M HCl. I have an Agilent 1100 HPLC system (quad pump, degasser, autosampler, diode array detector), and fear that my system's components may not be suitable for handling this relatively concentrated acid. Is this a reasonable concern, or are standard Agilent 1100's happy to handle HCl solutions? (Of course, I would thoroughly purge the HCl from the system after washing the column, so that HCl isn't hanging around to corrode parts over the long-term.) Thanks for your HPLC wisdom!
-
- Posts: 36
- Joined: Wed Jul 31, 2013 9:29 am
If the user manuals do not explicitly state that you can use that cleaning procedure, don't do it.Wash your columns on a separate system(we use pumps from retired systems).
-
- Posts: 252
- Joined: Sat Nov 07, 2009 6:27 pm
I sure as hell wouldn't introduce HCl into my 1100. The only thing more corrosive than the chloride ion is the chloride ion under acidic conditions.
-
- Posts: 269
- Joined: Fri Jun 29, 2007 1:48 pm
Many years ago I worked for a contract lab where we had a product up on stability and the test method used for either the impurities/assay used about 1M HCl in the mobile phase.
I can't tell you how nervous we were with running this method. We dedicated one of our 1100's to this method. We would use that 1100 for other methods, but we would only use the 1M HCl method on that particular 1100 HPLC.
Part of the procedure was to thoroughly passivate the system first before every run that was made.
We ran that stability method for year and a half to two years... never had any unusual problems with the 1100 during that time.
It's not something you really want to do, but if you passivate the system every time you do it, you might come out okay in the end.
I can't tell you how nervous we were with running this method. We dedicated one of our 1100's to this method. We would use that 1100 for other methods, but we would only use the 1M HCl method on that particular 1100 HPLC.
Part of the procedure was to thoroughly passivate the system first before every run that was made.
We ran that stability method for year and a half to two years... never had any unusual problems with the 1100 during that time.
It's not something you really want to do, but if you passivate the system every time you do it, you might come out okay in the end.
4 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 2 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 29 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 29 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science