Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hello everyone,

First, do we have to turn off GC everyday before we leave the lab? Do we just create a "sleeping" method with low gas flow rate and temperature when we are not using it? Is it safe to leave gases like hydrogen or oxygen flowing over-night without monitoring?

If we do want to turn off the GC, what steps should I follow?

Thanks,
Nina
GCs can safely be left overnight at whatever settings they are on during the day - thousands of them all over the world run automatically overnight. If you do want to put one to sleep overnight just reduce the oven temperature to a bit above room temp, leave the detectors and injectors at working temp, shut off the detector gasses (for ECD leave the make-up on), and use the gas saver feature to set a lower (20 - 30 ml/min) flow through the inlet.

Peter
Peter Apps
First, do we have to turn off GC everyday before we leave the lab?Is it safe to leave gases like hydrogen or oxygen flowing over-night without monitoring?
1st - definitely nope.
As for hydrogen. Read the manual. If it warnes not to leave FID unatended then you shouldn't - there will be no excuse in the worst case, you should be aware of every precautions from manual. But finally the decision is yours. You have to consider if it would be possible that hydrogen concentration might reach LEL in case of leak taking into account cubature of the room, ventillation etc.

If we do want to turn off the GC, what steps should I follow?
Cool down heating zones.
GCs can safely be left overnight at whatever settings they are on during the day - thousands of them all over the world run automatically overnight. If you do want to put one to sleep overnight just reduce the oven temperature to a bit above room temp, leave the detectors and injectors at working temp, shut off the detector gasses (for ECD leave the make-up on), and use the gas saver feature to set a lower (20 - 30 ml/min) flow through the inlet.

Peter
Thanks, your answer helps a lot.
Nina
4 posts Page 1 of 1

Who is online

In total there are 53 users online :: 0 registered, 0 hidden and 53 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 53 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry