*may we presume you got it to light just fine, if so, how'd you do it?Hi, I work with a Perkin Elmer Clarus 480, with FID, i repleace the Air tank and then the detector doesn´t ingite, but i can see the glow. I checked the flows and there were fine. Help, please!!!!
Around here, air and nitrogen have mutually incompatible connections to prevent just that (after some hospital patients in SA were given pure nitrogen instead of air).My guess is that the the 'Air' tank replaced was not an 'Air' tank but a Nitrogen tank...................