I inject an analyte solution(Dissolve in mobile phase) in HPLC, which elutes at 2.8 min (tr); however the solvents(methanol and acetonitrile) also have the absorption peak at 4.5min, the same intensity for each injection. Does it mean that the analyte has no interation with stationary phase in the C18 column? The lineary and standard curve of the analyte meet the HPLC method validation criteria. Can I still use the method to assay the concentration of the analyte?
The HPLC condition is shown as following, RP-HPLC
C18 column,
mobile phase: Acetonitrile: buffer(TFA ph=3.0): methanol = 12:40:48;
Flow rate: 1ml/min;
Waiting for some response! Thanks a lot!!!!