Hi all

I am intending to develop a PMS (Schedule, or Calender) for the GCs mainly and other equipment being used in our lab.

any recommendation

regards,
Hi all

I am intending to develop a PMS (Schedule, or Calender) for the GCs mainly and other equipment being used in our lab.

any recommendation

regards,
Preventative Maintenance will vary greatly depending on what type and how many samples you run each day.

We run 50-75 injections each day and over night so we pretty much change the lnserts daily. When I ran 8270 GC/MS we changed them daily along with clipping about 4 inches of column daily. Septa are usually good for 100-300 injections depending on temperature of the injection port.

If you have a log book from the last year where you preformed maintenance as needed, look through that and get the average time between maintenance and take about 70% of that and make it your schedule. That should cover most problems you would normally encounter before they happen.
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there are 50 users online :: 0 registered, 0 hidden and 50 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 50 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry