Pressure issues with Acetonitrile

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

10 posts Page 1 of 1
Morning,

Lately during my runs I have been seeing a massive spike in pressure when I flush the column with 100%ACN. Normally when my mobile phase is running (10%ACN/90%H2O) I have a backpressure of around 190-200bars but when I switch to 100%ACN the back pressure spikes to about 275-280bar and remains there until I revert back to the normal phase. Obviously this should not be happening but I do not know what could be causing it. Thoughts? Thanks

B
Do not connect the column outlet to the detector , and reverse flush the column.
Check if the pressure drops.

What is the column dimensions ? What is the column phase ( C-18 ? )
Column is an Agilent Poroshell120: 100mmx4.6mm, 2.7um particle size, C18

When I flip the column and run 100%ACN it is about half the pressure, 140bar. I flushed for an hour with a very minimum decrease in pressure, went from 140bar to about 138-139bar. When I put the column back in the correct direction the pressure is again at 280bar.
Possible causes are :
Blocked inlet frit of column ( which permits one way flow ).
The flow passage is partially blocked in the column head at the inlet side.

Also , you may cut an inch of the connection tubing ( if it is a PEEK tubing ) to the column , if there is a blockage in the tubing .
To remove any impurities blocking the column follow these steps :
1) separate the column from the outlet frit .
2) make the column drainage in a glass beaker and then wash the column with 60% methanol for about 60 min .
3) invert the column and repeat step (2) .

ph/Amr Tarek
Instrumental Unit Head at Otsuka Pharmaceutical Company
From the product manual:

Cleaning Your Column/Extending Column Life

For columns that can be backflushed (Poroshell 120, ZORBAX columns
with particles >1.8 µm and all Pursuit and Polaris columns), start with a
stronger (less polar) solvent.

1. Disconnect column from detector and run wash solvents into a beaker.
2. Start with your mobile phase without buffer salts (water/organic).
Run at 10 to 20 column volumes through.
3. Next, use 100% organic (methanol or acetonitrile).
4. Check pressure to see if it has returned to normal; if not, then,
5. Discard column or consider stronger conditions,
e.g., 75% acetontrile/25% isopropanol
6. Increase to 100% isopropanol, 100% methylene chloride or 100%
hexane (if you use methylene chloride or hexane, you will need to flush
the column with isopropanol prior to use and before returning to your
reversed-phase mobile phase).
Time flies like an arrow. Fruit flies like a banana.
I tried the 60% Methanol flush and that improves the pressure slightly. I've gone through the entire column flush procedure and that only provides minimal improvement as well. I just didn't know if there was something I was missing or not trying but I think with the content of the samples I'm running this will be a constant problem for me. Thanks for all your help!
As I mentioned earlier , it seems it is not a chemical contamination problem , so washing will not improve it.
It is a physical blockage somewhere and you have to locate it to get rid off.

If it is an inlet frit problem , you may try changing it but there is a risk of disturbing the packing.

You may try also remove and exchange the end-fittings ( inlet to outlet ) of the column to locate the problem.
I'm sorry I forgot to address that... I do not believe its a physical block in the system. When I place a new column on there is no issue in regards to the pressure. The problem only arises after I've run 50-60 samples.
I'm sorry I forgot to address that... I do not believe its a physical block in the system. When I place a new column on there is no issue in regards to the pressure. The problem only arises after I've run 50-60 samples.
That can be a problem with certain types of samples. It would be best to use a small guard column with prefilter. This would catch any contaminates that clog the main column and is much cheaper to replace than the main analytical column.

Guard Cartridge Holder

http://www.restek.com/catalog/view/3081

Guard Cartridge

http://www.restek.com/catalog/view/7005

Filter Frits

http://www.restek.com/catalog/view/920

These are what we use for analysis where dirty samples are a problem. The holder with filter frit has replaceable frits that are easy to change. Saves us a lot on column replacement when doing things like analyzing for Formaldehyde and PAH in tobacco samples.
The past is there to guide us into the future, not to dwell in.
10 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry