sidz28 » Fri Jan 03, 2014 7:11 pm
Hello, JGK.
Full disclaimer - I work for EST in technical support. I'm not sure if your difficulties have been with us, or with the service group that covers your territory for us, if you aren't located in the US. If it is with us, I sincerely apologize.
Reading through your comments, it seems to me that the 8100 is delivering the same amount of surrogate compound every sample. How are you measuring the amount? Is it quantified with an MS and an internal standard compound, or are you going by area count of the peak? What type of detector are you using if you're not using a mass spec?
It's an odd presentation, as the IS addition of the 8100 is a fixed-loop system similar to what Bigbear described. The 8100 is basically an 81-position version of the Varian Archon and OI 4552, so anyone familiar with those systems IS valves will know that you have a 1uL (or in some cases a 0.5uL) notch in the valve rotor that will fill with solution, then rotate back to be in line with your sample as it's delivered to the concentrator.
So for you to see that same amount of standard added for every run (and matching up with hand spiked amounts) leads me to believe that the standard addition of the 8100 is OK. Unless I'm misunderstanding, it sounds like the problem is that you recover twice as much as you expect of your spiked target analytes. Is that correct? I can't think of a logical explanation for that. Theoretically, I guess you can add twice the volume of sample, but you should notice that in the remaining volume in the vial.
Can you give a more thorough explanation as to what you're seeing? Specifically how you're measuring responses (detector type, measuring in qualitative area counts or quantified concentrations with an internal standard)?
If you prefer, you can email EST support directly at
support@estanalytical.com. We can do our best to help remotely.
Mike