 - Posts: 164
 - Joined: Wed Jul 03, 2013 1:51 am
 
we are experiencing quite intense interference from a contaminant at m/z 188.123 (See attached picture!). I was able to assign an elemental composition and I think it is diethylethanolamine. I changed the mobile phases and purged the HPLC system (using bottled HPLC water this time instead of MilliQ). That did not help at all.
What would you try cleaning next and has anyone experience this problem? I will changed the solvent lines next.Could it come from the air? The room where this MS is located does not have the greatest ventilation. We have always seen this ion but formerly at about 1% of what we see now.
Solvents are A: H2O (HPLC grade) and 0.1% formic acid from ampule. B: methanol (HPLC grade) and 0.1% formic acid from ampule.
Link to figure at higher resolution: http://i.imgur.com/jscZyPv.png
Thank you so much for your input.
Arne
					