Any ideas on why I get two different results for S/N ASTM depending on if I integrate only the peak of interest, or the peak of interest and a previous peak? As ever, the EZ Chrom help is of no use whatsoever
If you integrate only the peak of interest, everything else that's near this peak will be seen as noise by EzChrom. If there are other peaks present, no matter if they are system peaks, injection noise, gradient humps or "real" analyte peaks, is interpeted as noise so your s/n ratio will be really small. If you integrate these peaks, they are not considered noise anymore for EzChrom, so s/n gets higher.
AFAIK the EzChrom help explains which area of the chromatogram around the peak of interest are used for noise calculation. The explanation above of course only applies to peaks found in this area.
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 2 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry