Agilent GCMSD 6890/5973A and GC ALS Configuration

Posted: Tue Dec 17, 2013 6:37 am
by rochelle.m
Hi,

I would like to ask the correct procedure on how to configure GC ALS to GCMS. I tried to configure the instrument using "Config" then add new instrument. I write the GPIB address of GC and MSD but when I started up the unit and toggled down the Inlet/Injection Types, I cannot see GC ALS? I am using G1512A Autosampler. Am I missing something? Please help.

Re: Agilent GCMSD 6890/5973A and GC ALS Configuration

Posted: Tue Mar 18, 2014 9:20 pm
by AIM
Do you have the HPIB and remote cable going from the G1512A?

Re: Agilent GCMSD 6890/5973A and GC ALS Configuration

Posted: Wed Mar 19, 2014 12:12 am
by Consumer Products Guy
Are you trying to add the ALS to existing instrument? Then you don't "Add Instrument", you just change the configuration of your existing instrument in Configuration Editor.

Yes, get all your cables plugged in.