Hi,

I would like to ask the correct procedure on how to configure GC ALS to GCMS. I tried to configure the instrument using "Config" then add new instrument. I write the GPIB address of GC and MSD but when I started up the unit and toggled down the Inlet/Injection Types, I cannot see GC ALS? I am using G1512A Autosampler. Am I missing something? Please help.
Do you have the HPIB and remote cable going from the G1512A?
Best Regards,
Alex Hummell

Analytical Instrument Management
ph. 303.730.4136 ext. 902
cell. 720.203.5224
http://www.AIManalytical.com
Are you trying to add the ALS to existing instrument? Then you don't "Add Instrument", you just change the configuration of your existing instrument in Configuration Editor.

Yes, get all your cables plugged in.
3 posts Page 1 of 1

Who is online

In total there are 8 users online :: 1 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry