I am in complete agreement with Don; do not turn off the helium. Do a simple calculation; what is the cost of the volume of helium that you will use on standby in three days ? Compare that to the cost of a column, the cost in time of the extra cycles of conditioning that the column will need before you can run samples and the cost in time of the extra pump down on the MS to get rid of the air and moisture stuck to its internal surfaces.
Peter
Thanks,
Peter. Point taken, the gas will stay on. Can I at least move to a splitless flow, when not in use?
I am normally not happy with N2 or H2O above 5%. I normally keep it below 3% with the oxygen to nitrogen ratio indicating no leaks ( 28:32 should be 4:1 if air is getting in). If it goes higher than 3% I usually break out the 50% IPA/water and use that on all the lines and look for bubbling and a can of compressed air and look for M/Z 52 on the seals and connection on the instrument itself .
5% compared to peak 69 of cal gas, right? I'm afraid I don't quite get what means to "break out the 50% IPA/water and use that on all the lines". Our installing tech was adamant never to use Snoope or similar liquids to check for leaks in the GC/MS pipeworks.
For an extended idle period you might want to set your column flow to 0.5-0.3ml/min if the pressure controller can handle that low of a flow, with the column oven set about 105C to keep moisture from gathering on the column. My instruments only sit idle from when the Friday run ends sometime on Saturday until we start again on Monday, so my idle temp is 150C for the oven just to make sure any trace hydrocarbons don't hang around and any high boilers cook off over time, since these are purge and trap instruments the high boilers really aren't that high boiling.
On our GCs we normally change the septa ever few days. I did manage to get 300 injections on a septa once without any leaks but you have to keep the pressure from the nut very low so not to compress the septa and yet high enough to make it seal. Also keeping it cool extends the life, so if you are running a fairly high temp injector change them often.
The Dustoff works great for finding leaks, try it in the oven on the MS interface nut to make sure there are no leaks there, then back up through the system to make sure nothing comes in anywhere before that. You can also use it around the analyzer seals to check for leaks.
Invaluable informations,
James, I owe you my first born. A rather nosy pre-teen, btw.

I think -based on what you say- I can devise a safe idle method, that can move the flow through the column down to 0.2 mL. What about the temp in the injection port? You suggest that low temp extend its life. Can I keep it safely at 105 when not in use?
2 more questions. There is a purge flow setting in the injection port settings in the method that can also be set as low as 0.5 mL, default set at 5mL. What would you suggest about it, on idle? Secondly, the Dust Off was also suggested by our tech, what peaks should I look for? Some websites suggest a value of 62 for dichloroethane, is that correct?
Gabriel