UV acceptance criteria (limits) for Accuracy parameter

Basic questions from students; resources for projects and reports.

3 posts Page 1 of 1
Hi!
Can anyone guide me on the acceptance criteria (limits) for Accuracy parameter in method validation for a formulation containing 0.1% of drug by UV-Visible Spectrophotometer

-P.Sai Praveen
FDA ORA guidelines are 95.0 to 105.0% accuracy.

My question is how you are going to demonstrate that your assay is stability-indicating, as a degradation product could still contribute UV absorbance.

Ask about specificity as well.
Hi!
Could you please provide me the link for the FDA ORA guideline for analytical validation