Advertisement

UV acceptance criteria (limits) for Accuracy parameter

Basic questions from students; resources for projects and reports.

3 posts Page 1 of 1
Hi!
Can anyone guide me on the acceptance criteria (limits) for Accuracy parameter in method validation for a formulation containing 0.1% of drug by UV-Visible Spectrophotometer

-P.Sai Praveen
FDA ORA guidelines are 95.0 to 105.0% accuracy.

My question is how you are going to demonstrate that your assay is stability-indicating, as a degradation product could still contribute UV absorbance.

Ask about specificity as well.
Hi!
Could you please provide me the link for the FDA ORA guideline for analytical validation
3 posts Page 1 of 1

Who is online

In total there are 163 users online :: 0 registered, 0 hidden and 163 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am

Users browsing this forum: No registered users and 163 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry