I never worked with 6890 series with the packed column before.
However, what i know is that the 6890N series require a packed port which only support 1/8" packed column, and requires an additional 1/4" adaptor (~170US dollars) to work with my column. The 1/4" adapter does not have any glass insert, it combines both functions as a sleeve and an insert, which creates huge dead volumn.
The worst is the detector assembly. The TCD detector requires an additional adapter to work with my 1/4" column. However, unlike the S/SL injection capillary system, the detector adapter is loosely assembled to fit the whole apparatus. I will say, approximately 4 mL of empty space before the analytes reach the detector.
My result is acceptable, since we are using "loosely" defined acceptance criteria. The peak shape is broad, the baseline response is much higher compared to the same system using capillary system.
I can't seem to identify the problem source, maybe it is due to the copper-made 1/4" injection port adapter or the stainless steel 1/4" detector adapter.
What do you suggest to isolate and identify where the problems actually come from???