Any good information would be helpful.
Regards,
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
Running air samples you shouldn't make the source dirty very quickly at all. I run about 40-50 drinking waters samples each day by purge and trap and I go at least 6 months sometimes 18 months between source cleanings and even then they are not very dirty. If you are getting air in the analyzer when it is at high temp though it will oxidize the surface of the source some, which usually turns it a bluish or gold color and that needs to be cleaned off with some abrasive.we have a 5975 triple axis.
I found we had a small leak in our press connector. And a small leak on the autosampler.
The interface nut constantly loosens. Thanks for the suggestion for the oven I will see if they can preshrink several ferrules.
However, we have air samples and only 10 ml aliquot of a max of 200ppbv standard with a mass range from 6ngs to 20ngs per a run. We have just received the larger Draw out plates.
I will ask the floor to try it and see if it works.
For the ETP, any recommendations? I will see if i can order one to see if that helps with some of our problematic instruments.
Tomorrow, I will the runs and see if they are still having response climb during the day. We run about 60 samples a day.