Hi all,

I have an application that requires indirect UV of 280 nm. I am not so sure how to setup the parameters in the DAD: the wavelength, the bandwidth, the ref...
Could anyone have done this and help.
Thanks.
Besides too little information:

Just use standard settings for the UV detector. What you need to change is the polarity of the signal. Or your software needs to be able to cope with negative signals.
In case of a signal cable connected to a integration system, interchanging of the two signal cables may help.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Besides too little information:

Just use standard settings for the UV detector. What you need to change is the polarity of the signal. Or your software needs to be able to cope with negative signals.
In case of a signal cable connected to a integration system, interchanging of the two signal cables may help.
Thanks for your input. I just set the ref wavelength at 280nm and the signal is 360nm. It works.
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry