The HPLC parameters are as follows:
Mobile Phase Preparation: (A) Dissolve 5.5 grams of Na 1-Heptanesulfonate in 50 mL of H2O, then bring to volume with GAA in a 100 mL VF. In a 1 L bottle add 40 mL of the Na 1-Heptanesulfonate solution + 4 mL TEA + 960 mL H2O, pH 3.2. (B) Use 100% ACN
Gradient: 0 - 5 min 50% B; 5.5 - 6.5 min decrease MP B to 20%; 6.5 - 8 min ramp MP B back to 50%, hold for 8 min to re-equilibrate.
Column Temp = 45 C; Injection Volume = 10 uL; Wavelength = 254 nm; Diluent = MeOH
Column: Agilent Poroshell EC-C8, 2.7 um, 150 x 4.6 mm.
The issue I have ran into is during initial method development and running my stressed samples my Retention time was 7.7 mins, when I ran my next sequence my retention time shifted to 6.8 mins (keep in mind this is the same column and mobile phases from the same bottles). I thought the column was going bad so I purchases a new one. When I used the new column with the same mobile phase as used above the retention time was 6.3 minutes, when I went back to the previous column it also shifted forward to 6.3 minutes.
I have determined to separate the placebo peaks from the API I need to use a C8 rather than a C18, and that I obtain the best peak shape with a smaller particle size column.
Any thoughts as to how I should proceed or what may have caused the huge retention time shift would be greatly appreciated.