multiple headspace extractions - calculations

Discussions about sample preparation: extraction, cleanup, derivatization, etc.

2 posts Page 1 of 1
I have a question regarding multiple headspace extractions.

After plotting the natural log of peak areas, you obtain the slope of the linear fit from this.

Is there a general lower limit for an acceptable slope? A colleague mentioned to try to get the slope above 0.4. I have seen some technical papers with slopes as low as 0.05! I am looking at high boilers in elastomers, and in order to increase the slope, the temperature has to be increased too high and lots of degradation peaks are observed.
Obviously the steeper the slope the better, but what really matters is the uncertainty of the slope compared to its steepness.

Flat slopes mean that the quantity of what you are measuring is not being depleted enough by the headspace sampling on each run. If you reduce the amount of sample in the vial you should see steeper slopes.

Peter
Peter Apps
2 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry