Issue with flows - HP5890 series II

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi everyone,
I'm trying to setup my old cromatograph. It's start up normally. The head pressure is ok. But I dont have all flows in the detector (FID). The air flow works properly, but the h2 and n2 dont appears (sometimes, the h2 work a few time). Anyone can help me?
Hi

Initially check that you have hydrogen entering the 5890. Secondly check the jet is not blocked, the hydrogen goes through the jet.

This accounts for most issues, get back to us if this does or doesn't work.
Hi

Initially check that you have hydrogen entering the 5890. Secondly check the jet is not blocked, the hydrogen goes through the jet.

This accounts for most issues, get back to us if this does or doesn't work.
Thanks for answer. The jet was clean. I found the problem. Was the broken liner. It was a disruption. Now, the HP5890 it's new :D
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 1 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry