Dear all,

I am trying to learn more about calibration of ESI instruments and i am confused and stuck on a few facts.

From a book called Analytical Method Validation and Instrument Performance Verification, Edited by Chung Chow Chan, Herman Lam, Y. C. Lee, and Xue-Ming Zhang.

It writes that "The right calibrants for LC-ESI-MS should (1) not give memory effects; (2) not cause source contamination through the introduction of nonvolatile material; (3) be applicable in both positive- and negative-ion mode. "

However what is the difference between memory effects and source contamination? From what i have gathered source contamination is one of the reasons of memory effect and they should be the same am i right to say so?
Absolutely !!

"Memory effect" is just a hand-waving, catch-all description of contamination from previous analyses.

Injection port/sample delivery lines/ionization source etc. are all possible sites of contamination.
Hi JMB,

Thanks for helping me out . For cases where the calibrant is "sticky" and difficult to remove from the column / require more than reasonable amount of rinse time, is there a term for such calibrants?

If i rephrase the statement to say "The right calibrants for LC-ESI-MS should (1) not be "sticky" resulting in memory effects; (2) not cause source contamination through the introduction of nonvolatile material; (3) be applicable in both positive- and negative-ion mode."

Would it be more correct?
Yes, I think that your rewrite is acceptable, but if you keep your phrasing I would insert a comma ........"sticky",.....

You might consider rephrasing to something similar to

........should (1) be easily washed from the system, otherwise memory effects may result, (2)....

which avoids the use of the term "sticky".

Another requirement for mass calibrants is that they should also be readily available in a high state of purity; this usually implies a crystalline solid.

As a practical point, mass calibrants are usually injected downstream of the HPLC column via an injection port.
4 posts Page 1 of 1

Who is online

In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry