S/N as part of performing bi-annual calibration

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
All,
I am curious to know if people perform a signal to noise check as part of a bi-annual or annual system check/calibration. I know noise and drift of a detector check is normally performed but do people check S/N as per USP with a known peak? I'm used to seeing S/N being performed on a method to method basis as part of a system suitability requirement run to run but have not seen it in a requirement to be performed as part of an LC calibration.
If you wanted to do it why not repeat the S/N test that was done on the install qualification (assuming you work in a regulated environment).

We don't tend to do this test as a part of our yearly calibration.

As you allude to, this test characteristic can be more picked up quite readily from a system suitability requirement every time you run a suitable method.
2 posts Page 1 of 1

Who is online

In total there are 25 users online :: 2 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 23 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry