No signal on LCQ Advantage

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hey there,

I get no signal. Running all the diagnostic tests the Trap Offset and the Intermultipole Lens failed. Any ideas?

Thanks a lot
given the age of the system, there is a very real risk that this is a failure of some board inside. If you're in Europe, you should definitely seek advice from Thermo. They will almost certainly be very helpful about advising on likely causes without cost, but if it is a failed board, unfortunately you may be looking at a fairly expensive repair.
Thanks for your advice. After shutting down the system everything seemed all right. no problems with trap offset etc. I got good signal. But after some measurements it came back. No signal etc. We´ll ask Thermo...
Could be a serious case of charging.. Might require a good thorough clean.
Kind regards
Leadazide
4 posts Page 1 of 1

Who is online

In total there are 38 users online :: 0 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry