Can anyone suggest a reason for variations seen in manual injections of standard directly into a GC injection port? I am working with an Agilent 6890N GC/5973 Inert GC. Inj port temp = 180 C; split 30:1; 1 uL inj vol. There are no leaks anywhere in the system. Thanks.
Poor syringe handling is by far the most likely.

Peter
Peter Apps
What is the solvent for your samples?
Poor syringe handling is by far the most likely.

Peter
The time the needle is in the injection port can make a difference in your analyte response. Also the needle depth at time of injection or the speed you inject. It takes a lot of practice to be reproducible with manual injection techniques, but it can be done.
The past is there to guide us into the future, not to dwell in.
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry