cleaning FID
Discussions about GC and other "gas phase" separation techniques.
5 posts
Page 1 of 1
does it necessary cleaning the collector ?
saba
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
If a FID is dirty and contaminated noise will be generated in the signal.
Usually, deposits from silylating reagents are the problem. Most other chemicals eluting from the column burn and leave little or no deposits.
Noise is also generated by contaminated gas, carrier, fuel, or air. Use high quality clean gases. Also be sure the cables are not frayed, loose, or corroded at the detector or the amplifier ends.
Usually, deposits from silylating reagents are the problem. Most other chemicals eluting from the column burn and leave little or no deposits.
Noise is also generated by contaminated gas, carrier, fuel, or air. Use high quality clean gases. Also be sure the cables are not frayed, loose, or corroded at the detector or the amplifier ends.
-
- Posts: 34
- Joined: Thu Jul 28, 2005 8:06 pm
I will say that you clean it only when your experiments didn't generate satisfactory results due to possible performance defects in detector.does it necessary cleaning the collector ?
usually, the white residue is the result of column bleed. the residue generates a lot of noise in your baseline and the FID might require some cleaning. The black residue is usually the result of your analytes, and it doesn't affect as much as the white residue.
Thanks and Best Wishes
-
- Posts: 26
- Joined: Wed Sep 08, 2004 6:11 pm
I run a simulated distillation of lube base oils. The samples are diluted in CS2. The sulfur deposits on the chimney of the FID. It has a fairly nice yellow tint. Since the column is a wide bore 0.53u, the jet is the small orfice for packed columns. This jet becomes clogged with white residue. I can actually look down the chimney and see the flame burning, which normally you can't see. Eventually, the signal will become noisey and sometimes drop out to nothing. Check the bottom seal and replace if hard and cracked.
-
- Posts: 28
- Joined: Sun Aug 07, 2005 11:06 am
thanks for explaination.
saba
5 posts
Page 1 of 1
Who is online
In total there are 52 users online :: 1 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 51 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 51 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science