Weird baseline

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Does anyone know what might be causing this? the baseline is steady then it drops, then it's steady, then it drops again.

Image

presure is steady with no significant changes.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
Lamp flicker?
Small air bubble accumulating in the cell gradually and then suddenly breaking loose?
Dirty cell windows?

If this were my problem, I'd check the noise with no flow. If it gets better, that eliminates the lamp. Setting a reference wavelength would diagnose lamp flicker (should get better with reference on).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
I turned off the flow as you suggested and yes, it appeared to be the lamp.

The lamp has been replaced an that baseline behavior is gone. Now the baseline is drifting upwards slightly throughout the sequence. I can still integrate my peaks, however.

Thanks, Tom.
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island
3 posts Page 1 of 1

Who is online

In total there are 374 users online :: 2 registered, 0 hidden and 372 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Majestic-12 [Bot] and 372 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry