Gas analysis

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
I have Trace GC with FID detector, capillary column, where we want to start to measure gas, methane, through the injector in split-less mode using gas-tight syringe. Is this going to work or am i force to use a loop injector? it's for the injection of 100ul calibration gas cylinder.
Have you tried it? Is methane the only thing in your mixture? There's probably no faster GC analysis than methane. You could probably make 10 injections in 30 minutes. Methane is pretty light so it might work for that. Condensation of heavier materials can manifest itself in an apparent lack of precision.

I've always had trouble getting good precision for direct injection of gas samples like you describe. The only way I've ever been able to get the precision I want/need is to use a fixed loop (inside the oven, 40 °C). Even then, you need to be sure that you're making the injection of a sample at atmospheric pressure.
2 posts Page 1 of 1

Who is online

In total there are 54 users online :: 2 registered, 0 hidden and 52 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot], Semrush [Bot] and 52 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry