correct procedure to prepare a phoshate buffer at pH=7

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Disovle 3.4 g KH2PO4 (potassium phosphate monobasic) in 1 L water, pH of the solution is around pH=4.5. Then use KOH solution to adjust pH to pH=7. then you get a phosphate buffer that buffers around pH=7.

Is that a correct procedure to prepare buffer solution?
Or, you could simply do it by weight. Here's an on-line calculator for sodium phosphate:
http://www.egr.msu.edu/biofuelcell/tool ... phate.html

Just scale those by the relative molecular weights of the potassium and sodium salts:
monobasic: 136/120
dibasic: 174/142

No pH meter needed.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Tom's approach is preferable to me as it is easier to duplicate mass determination(s) for salts than it is to titer a solution containing a single salt due to differences in pH meters and other variables in preparation.
MattM
Dear Matt

The titration approsch is also no problem in this case. The second pKa of phosphoric acid is 7.2 therefore change in pH is slow and reaching the expected pH value is pretty straight forward. There is almost no danger of overshooting.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
Hi Markus,

I agree with you in that the titration approach works well when properly executed--with the caveat being the care with which multiple analysts calibrate/measure pH/take care of the pH electrodes. I've observed less troubles with weighing the buffer ingredients than with titration in method transfer exercises (this is just my own observation set where I have been). Of course, if everyone is on the same page with regard to measuring pH and the meters/electrodes are treated properly, either buffer preparation method will afford the proper final result.

I've also been in places where, when isocratic eluents containing organic/aqueous (buffer) solutions are pre-mixed off-line of a HPLC instrument, that the solutions were Weighed before mixing (instead of, say, using graduated cylinders) in an effort to ensure a higher degree of uniformity between different analysts and labs. Nothing like using a pycnometer!

All in the effort to maintain consistency...
MattM
5 posts Page 1 of 1

Who is online

In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Majestic-12 [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry