what is the appropriate shut down method for GCMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Need to standby over weekend. what it the gc/ms shut down method? Carrier gas reduced to 0.5ml/min? Inlet temperature 150C, split mode 20:1 oven 150C, transfer line temp 250C, MS QUAD and source 150C and 230C? Anything that I need to change?
Need to standby over weekend. what it the gc/ms shut down method? Carrier gas reduced to 0.5ml/min? Inlet temperature 150C, split mode 20:1 oven 150C, transfer line temp 250C, MS QUAD and source 150C and 230C? Anything that I need to change?
Over a weekend we don't change anything except keep the oven at least 150c instead of the normal starting temp of around 50c. That will help prevent any hydrocarbons that might be present from accumulating on the column phase. You may lower the split if you are worried about using too much helium, or even plumb in a nitrogen source if really worried about wasting helium, but we just count it as an operating expense. Quad and source temps can remain at running levels, it doesn't take that much more electricity to keep them heated, and you don't have to wait for them to come back to temp on Monday to start analysis.

The old HP5970 had a setting to reduce the turbo pump down to 80% speed as a standby, but keeping them at 100% seemed to make them last longer.

Unless you are shutting down for several months I really don't see any advantage to changing anything over what you would for an overnight pause in analysis.
The past is there to guide us into the future, not to dwell in.
Thank you so much!
Need to standby over weekend. what it the gc/ms shut down method? Carrier gas reduced to 0.5ml/min? Inlet temperature 150C, split mode 20:1 oven 150C, transfer line temp 250C, MS QUAD and source 150C and 230C? Anything that I need to change?
Over a weekend we don't change anything except keep the oven at least 150c instead of the normal starting temp of around 50c. That will help prevent any hydrocarbons that might be present from accumulating on the column phase. You may lower the split if you are worried about using too much helium, or even plumb in a nitrogen source if really worried about wasting helium, but we just count it as an operating expense. Quad and source temps can remain at running levels, it doesn't take that much more electricity to keep them heated, and you don't have to wait for them to come back to temp on Monday to start analysis.

The old HP5970 had a setting to reduce the turbo pump down to 80% speed as a standby, but keeping them at 100% seemed to make them last longer.

Unless you are shutting down for several months I really don't see any advantage to changing anything over what you would for an overnight pause in analysis.
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry