what is the appropriate shut down method for GCMS
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
Need to standby over weekend. what it the gc/ms shut down method? Carrier gas reduced to 0.5ml/min? Inlet temperature 150C, split mode 20:1 oven 150C, transfer line temp 250C, MS QUAD and source 150C and 230C? Anything that I need to change?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Over a weekend we don't change anything except keep the oven at least 150c instead of the normal starting temp of around 50c. That will help prevent any hydrocarbons that might be present from accumulating on the column phase. You may lower the split if you are worried about using too much helium, or even plumb in a nitrogen source if really worried about wasting helium, but we just count it as an operating expense. Quad and source temps can remain at running levels, it doesn't take that much more electricity to keep them heated, and you don't have to wait for them to come back to temp on Monday to start analysis.Need to standby over weekend. what it the gc/ms shut down method? Carrier gas reduced to 0.5ml/min? Inlet temperature 150C, split mode 20:1 oven 150C, transfer line temp 250C, MS QUAD and source 150C and 230C? Anything that I need to change?
The old HP5970 had a setting to reduce the turbo pump down to 80% speed as a standby, but keeping them at 100% seemed to make them last longer.
Unless you are shutting down for several months I really don't see any advantage to changing anything over what you would for an overnight pause in analysis.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Thank you so much!
Over a weekend we don't change anything except keep the oven at least 150c instead of the normal starting temp of around 50c. That will help prevent any hydrocarbons that might be present from accumulating on the column phase. You may lower the split if you are worried about using too much helium, or even plumb in a nitrogen source if really worried about wasting helium, but we just count it as an operating expense. Quad and source temps can remain at running levels, it doesn't take that much more electricity to keep them heated, and you don't have to wait for them to come back to temp on Monday to start analysis.Need to standby over weekend. what it the gc/ms shut down method? Carrier gas reduced to 0.5ml/min? Inlet temperature 150C, split mode 20:1 oven 150C, transfer line temp 250C, MS QUAD and source 150C and 230C? Anything that I need to change?
The old HP5970 had a setting to reduce the turbo pump down to 80% speed as a standby, but keeping them at 100% seemed to make them last longer.
Unless you are shutting down for several months I really don't see any advantage to changing anything over what you would for an overnight pause in analysis.
3 posts
Page 1 of 1
Who is online
In total there are 139 users online :: 0 registered, 0 hidden and 139 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 139 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 139 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science