- Posts: 1
- Joined: Fri Sep 27, 2013 8:44 am
I have a special phenomenon on my TCD.
In front of the O2 peak the base line goes negative (negative area of ca. 200). The problem is that at this retention time I expect the H2 peak.
Because of that, it is problematic to detect low concentrations of H2 (e.g. 50ppm)
TCD chromatogram:
http://s14.directupload.net/file/d/3393 ... i8_jpg.htm
red: gas sample with 5000ppm H2, 25% O2, 72% N2 (H2 peak overlaps negative peak)
black: without sample injection (no negative peak)
blue: air injected (negative peak)
carrier gas: Argon 5.0
column: Mole Sieve 5A 1m, micropack
Is it possible to get a different separation of these three gases with another column?
Thanks and best regards
Thomas