So the question is to do the SS in order to validate the HPLC system before inject the samples or only for to find the precision of the method?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Wow, for me that's really a new way to interpret that SST criterion. Any decent auditor will kick them in the ***.For example, if in the SS for the 5 first std injections the RSD>2.0%, they can use the other std injections after the samples injections for to arrive RSD<2.0%.
Just my 2 cents:So, what do think about the fact that "that people" wants to eliminate one injection of the SS (considering the first 5) because of the RSD>2.0%. "That people" says that they have an internal SOP (named: treatment of outlier data) that could eliminate data when one injection area is over +/- 3 x stdev of the average data excluding the outlier data (or outlier area injection). I think is ridiculous but... we can eliminate one injection of the SS with an internal SOP? The internal SOP do not say anything about SS, is a general treatment of outlier data.
First, USP has harmonized their requirement with Ph Eur with regard to RSD for assays.La USP says to run 5 replicate injections. These injections must to be consecutive or not? I think the SS must to be run at the beginning in order to be sure that the system is OK or the system precision is OK (RSD < 2.0%). But some people says that these injections could not to be consecutive, it means that they can use std injections of the bracket (std's after samples injections) in order to arrive to RSD<2.0%. For example, if in the SS for the 5 first std injections the RSD>2.0%, they can use the other std injections after the samples injections for to arrive RSD<2.0%.
So the question is to do the SS in order to validate the HPLC system before inject the samples or only for to find the precision of the method?
They use 2 preparations, 2 injections for each preparation, it means 4 injections. If one of the 4 is outlier they eliminate 1 and work with only 3 areas.Just my 2 cents:So, what do think about the fact that "that people" wants to eliminate one injection of the SS (considering the first 5) because of the RSD>2.0%. "That people" says that they have an internal SOP (named: treatment of outlier data) that could eliminate data when one injection area is over +/- 3 x stdev of the average data excluding the outlier data (or outlier area injection). I think is ridiculous but... we can eliminate one injection of the SS with an internal SOP? The internal SOP do not say anything about SS, is a general treatment of outlier data.
- If the SST says %RSD from 6 injection <2.0%, you must consider 6 injections! There's NO excuse for omitting one. The SST is a vital part of each method. If it's not fulfilled, the system is not suitable. You're not allowed to circumvent any SST parameter with a general SOP!
- That "outlier SOP" is some sort of smart thing which can go terribly wrong! If they work in a regulated environment, there MUST be an investigation of outlier data. You cannot just eliminate any data because of a general SOP! Furthermore, if they use the RSD of the data to decide wether there's an outlier or not, there's one further problem: RSD starts to make sense only if you have at least 3 values. That means, including the possible outlier, you need at least 4 values. So they must generally do 4-fold injections of each sample if they want to omit one injection...
Summarized, any capable auditor will have much fun with that outlier SOP...
Consider this scenario:While I agree that doing the SS *before* running samples is a good idea, I believe that you *can* use the check samples for the SS. There are two major problems with that, however:
1. You don't know until the end of the run if you passed or failed SS. If you fail, then all the data obtained are deemed invalid and the results are not usable.
2. You have to decide ahead of time to do this; you can't change the rules in mid-stream. So if you failed an initial SS, you cannot just keep running check samples until you pass.
All things considered, using the SS as check samples is a bad idea.
This is a noble and honorable attitude, but I'm afraid that environmental concerns are not really a strong argument for auditors or inspectorsHi everybody,
I think this kind of SST is suitable for preparation of working standard only, It waste time and material as much as possible (toxic materials for environment!). When you use a calibrated system and have a good program for P.M. and intermediate checks on your HPLC Column, this process is not necessary as this type, I am agree that check RSD for 6 Injections of STD is enough.
Please be more tender with our environment for our future race.
Why for precision studies only 6 injections should be given? & for system Suitability only 5 injections ? What is the logic behind the digit 5 and 6?