Irritating contamination in the C20 - C40 range coming from our lab, in processed samples and blanks. I don't believe it's the instrument/gas/vials/solvent as air injections and unprocessed solvent injections are clean. Looks like fish-fin style polysiloxane septa contamination, but it is absolutely alkanes as confirmed by FID and GCMS methods. Dozens of tests inconclusive, as this stuff seems to be in everything, and is sporadic in concentration but getting worse. Has anybody seen this particular stuff before? It is provoking my ire.

Image

Thanks,
Stunt
We have had problems like that where a wash bottle in the prep area was contaminated with some oil from the air. It was in quite a few samples before it was found. I can't remember the source, but it seemed odd at the time.
We had the same problem. Have you tested your gloves? Your gloves your lab tech/chemist wear are in contact with everything. Thats what happened with us! false positives everywhere!. The main scource of the contamination for us was that technicans were folding filter paper with gloves on and that was the main scource!
Thanks seamoro/Steve. Yeah, we tested the hell out of the gloves, and it was there. Even when we were testing without using any gloves its there. It's really bugging me, and starting to impact quality.

Anybody else?
Do you by any chance have a vacuum pump whose outlet is venting to your lab air ? Or is anyone using house compressed air ?

Peter
Peter Apps
What are you doing your extractions with? Glass or plastic?
Stunt,

What are the samples?
What is the GC configuration, inlet, column?
This contamination appears in the first injection after a standby period?
7 posts Page 1 of 1

Who is online

In total there are 37 users online :: 0 registered, 0 hidden and 37 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 37 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry