one compound has too much fragmentation, so sensitivity is very low. How can I decrease the fragmentation and increase the sensitivity.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Extract an ion of interest and use that, or use SIM (selected ion monitoring).one compound has too much fragmentation, so sensitivity is very low. How can I decrease the fragmentation and increase the sensitivity.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
already used SIM.
Extract an ion of interest and use that, or use SIM (selected ion monitoring).one compound has too much fragmentation, so sensitivity is very low. How can I decrease the fragmentation and increase the sensitivity.
-
- Posts: 959
- Joined: Thu Apr 18, 2013 2:10 pm
To my knowledge, you (thankfully) have no control over how the mass spectrometer fragments the molecules coming out of the GC. If you had control over the amount/degree of fragmentation, the libraries we have for identification would be useless. It pretty much is what it is.
Going to something like CI might be an option for you?
What is the analyte?
Going to something like CI might be an option for you?
What is the analyte?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
One option on an Agilent 5973 or 5975 would be to lower the Emission setting which decreases the current (but not the Electron Volts) at the filament. You will sacrifice a little overall sensitivity but it will bias the fragmentation towards higher masses because of less total fragmentation. The ion fragments should be the same, just biased more to the heavier masses.
If it is a compound with multiple chlorine molecules, well there isn't much you can do in that case, all the permutations of having multiple chlorine molecules with 35 and 37 masses causes a lot of fragments and lowers the intensity of each.
If it is a compound with multiple chlorine molecules, well there isn't much you can do in that case, all the permutations of having multiple chlorine molecules with 35 and 37 masses causes a lot of fragments and lowers the intensity of each.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Thanks so much!
One option on an Agilent 5973 or 5975 would be to lower the Emission setting which decreases the current (but not the Electron Volts) at the filament. You will sacrifice a little overall sensitivity but it will bias the fragmentation towards higher masses because of less total fragmentation. The ion fragments should be the same, just biased more to the heavier masses.
If it is a compound with multiple chlorine molecules, well there isn't much you can do in that case, all the permutations of having multiple chlorine molecules with 35 and 37 masses causes a lot of fragments and lowers the intensity of each.
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
Some models of Agilent will allow to to lower the electron energy, even in EI mode (70eV), though this produces non standard EI spectra. A lower source temp may also reduce fragmentation a bit.
7 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Bing [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Bing [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science