the same samples were injected twice back to back. why the second injections always have around 30-40percent response factor increase (IS corrected). What could be the problems?
I do a lot of SPME and I do not detect this phenomenon. My precision is generally better than this.

What is the concentration of your analyte? 30% of 100 parts-per-trillion is only 70 or 130 parts-per-trillion in the reported result. Either way, it's a ridiculously small amount.
concentration is around 20ppb.
I do a lot of SPME and I do not detect this phenomenon. My precision is generally better than this.

What is the concentration of your analyte? 30% of 100 parts-per-trillion is only 70 or 130 parts-per-trillion in the reported result. Either way, it's a ridiculously small amount.
the same samples were injected twice back to back. why the second injections always have around 30-40percent response factor increase (IS corrected). What could be the problems?
Have you done a blank injection using the same fiber right after the first injection? If so does it indicate any carryover?
The past is there to guide us into the future, not to dwell in.
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
I might need to inject blank inbetween.
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
Hi Jame, I did not see carry over for the sample for blank injection. This is wired.
the same samples were injected twice back to back. why the second injections always have around 30-40percent response factor increase (IS corrected). What could be the problems?
Have you done a blank injection using the same fiber right after the first injection? If so does it indicate any carryover?
if this is the case, then why Internal standards in second injection always have lower peak area (almost decrease by half) than the first injection for all five standards. Analytes peak area remain very similar between two injections for the same samples.


On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
One thing I notice is that I have very noticeable column bleeding. Does that affect my reproducibility?

On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
One thing I notice is that I have very noticeable column bleeding. Does that affect my reproducibility?

On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
If you have air leaking at the injection port causing the column bleed then that could be causing problems.
The past is there to guide us into the future, not to dwell in.
i do not have leaking issue.
Actually, I solve the problem by decreasing the extraction time. probably it is the competitive adsorption.
One thing I notice is that I have very noticeable column bleeding. Does that affect my reproducibility?

On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
If you have air leaking at the injection port causing the column bleed then that could be causing problems.
11 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry