precision of SPME
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
11 posts
Page 1 of 1
the same samples were injected twice back to back. why the second injections always have around 30-40percent response factor increase (IS corrected). What could be the problems?
-
- Posts: 959
- Joined: Thu Apr 18, 2013 2:10 pm
I do a lot of SPME and I do not detect this phenomenon. My precision is generally better than this.
What is the concentration of your analyte? 30% of 100 parts-per-trillion is only 70 or 130 parts-per-trillion in the reported result. Either way, it's a ridiculously small amount.
What is the concentration of your analyte? 30% of 100 parts-per-trillion is only 70 or 130 parts-per-trillion in the reported result. Either way, it's a ridiculously small amount.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
concentration is around 20ppb.
I do a lot of SPME and I do not detect this phenomenon. My precision is generally better than this.
What is the concentration of your analyte? 30% of 100 parts-per-trillion is only 70 or 130 parts-per-trillion in the reported result. Either way, it's a ridiculously small amount.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Have you done a blank injection using the same fiber right after the first injection? If so does it indicate any carryover?the same samples were injected twice back to back. why the second injections always have around 30-40percent response factor increase (IS corrected). What could be the problems?
The past is there to guide us into the future, not to dwell in.
-
- Posts: 959
- Joined: Thu Apr 18, 2013 2:10 pm
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
I might need to inject blank inbetween.
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
Hi Jame, I did not see carry over for the sample for blank injection. This is wired.
Have you done a blank injection using the same fiber right after the first injection? If so does it indicate any carryover?the same samples were injected twice back to back. why the second injections always have around 30-40percent response factor increase (IS corrected). What could be the problems?
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
if this is the case, then why Internal standards in second injection always have lower peak area (almost decrease by half) than the first injection for all five standards. Analytes peak area remain very similar between two injections for the same samples.
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
One thing I notice is that I have very noticeable column bleeding. Does that affect my reproducibility?
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
If you have air leaking at the injection port causing the column bleed then that could be causing problems.One thing I notice is that I have very noticeable column bleeding. Does that affect my reproducibility?
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
The past is there to guide us into the future, not to dwell in.
-
- Posts: 206
- Joined: Sun Oct 09, 2011 5:08 pm
i do not have leaking issue.
Actually, I solve the problem by decreasing the extraction time. probably it is the competitive adsorption.
Actually, I solve the problem by decreasing the extraction time. probably it is the competitive adsorption.
If you have air leaking at the injection port causing the column bleed then that could be causing problems.One thing I notice is that I have very noticeable column bleeding. Does that affect my reproducibility?
On "typical" things that are 20 ppb, I see better precision than +/- 30%. I agree with James_Ball. Is there something that's in your blank that is causing you trouble? Something coming from your injection (a siloxane from the fiber or the inlet septum) system?
11 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science