agilent 6890 vs 7890 with 5975 MSD - running H2

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Looking to purchase a GCMS system - any advice on advantages/disadvantages on 6890/7890 for performance/maintenance/operations/capabilities etc. other than the obvious cost difference? Any advice on setting up a "new-used" system to run H2 as the carrier?

Thanks in advance!
Check with Agilent about H2 carrier gas with MS. They have specific parts requirements for the source.
The 5975 needs a drawout plate with a larger aperture for H2 and some 5975's have a magnet in the source with no serial number on it that needs to be replaced for H2. Other considerations are you should probably get a performance turbo to efficiently pump H2.

Any Agilent GC should do hydrogen fine even a 5890 not that you would consider one. Don't inject chlorinated solvent or you could get HCl gas in the system.
Thanks for the really helpful advice. After seeing these posts, I picked up a copy of some recommended Agilent procedures for running H2 carrier gas. For now, I think I will stick with He...
4 posts Page 1 of 1

Who is online

In total there are 47 users online :: 2 registered, 0 hidden and 45 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Semrush [Bot] and 45 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry