Hi All:

I'm interested in elevated temperature, ballistic gradient RP-HPLC.

My column is 2.1 x 33 mm, 5 micron (PS-DVB)
Temp: 80 C
Flow rate: 1 mL/min
A = 30 mM diethylamine, water
B = 30 mM diethylamine, MeCN
My gradient is 2 to 99% B over 1 min, hold 99% B for 1 min, return to 2% B over 0.1 min, followed by 3 min re-equilibration period. The delay volume for this system is approx 0.6 mL.

The mobile phase is not pre-heated. Instead, I route pump outlet into a column oven heat exchanger, like in this picture.

http://www.spectralabsci.com/images/HP- ... rtment.jpg

The retention times for replication runs are remarkably consistent. SD < 0.5%

I'm curious to ask,
1) how will I know if the incoming mobile phase is matched to the temperature of the column oven?

2) When I run a 5 minute (2 to 99% B) gradient at ambient temperature, I find retention times (neutral, small molecule compounds) are linearly related to those measured in a high-temp (80 C) ballistic gradient (2 to 99% in 1 minute.) Is this to be expected?

Again, assuming neutral, small molecule solutes, in the absence of stationary phase degradation (I'm using PS-DVB), under what circumstances could selectivity or elution order change?
Selectivity and elution order are not changed, your PEEK tubing and the heat exchanger are ok, I would not expect a temperature Gradient on the column.
Gerhard Kratz, Kratz_Gerhard@web.de
2 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry