First off, my GC/MS conditions:
He at flow rate of 1.8 ml/min
Cool on-column injection, inlet temp mirrors oven temp
DB5-MS, 30 m X 0.25mm
MS transfer line 280dC
EI mode with quad set to 150dC and source set to 230dC
I am having a problem that I come across more often than I want to, which causes dowtime on my instrument. When I have to change my carrier gas, after running the GC/MS system overnight I am still getting Air (actually N2, m/z 28) values at around 10% or even higher. Strangely, my water (m/z 18) value is well below 1% and the ratio of N2 to O2 is much higher than in atmospheric air. The new He tank had been unused prior to installation. I have also read and heard differing specifications. In the help guide, the spec for N2 and water are <5% and <10%, respectively. BUT in the tune evaluation, the N2 spec is <10%. Folks at Agilent tech services have also told me that N2 < 10% is acceptable. SO I am unsure which applies.
Interestingly, when I remove the septum inlet the N2 level falls to below 5%. I've gone thru the system from the He tank, copper tubing, molecular sieve, inlet, column, MS transfer inlet and can find no apparent leaks. I've sprayed DUST-OFF in the GC oven and around the MS door and did not see the corresponding ion for Difluoroethane, so a leak in the MS is unlikely. I've determined that the issue has to be on the GC side, but if there were a leak and the system were pulling atmospheric air, wouldn't the N2/O2 ratio reflect this?
Or does the system just need that much time to pump the air thru? It's been nearly 30 hours since I've changed the tank. Unless of course, the tank is bad...which would be a gamble every time I have to change carrier gas tanks.
Has anyone seen anything like this, and if so what to do about it?
CT