Reason for negative peak after main peak in RP HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Request for the reason .
Please supply more information such as column, mobile phase, detector type, flow rate, analyte in order to receive an explanation.
In addition to skunked_once please let us know if your sample is in the same solvent diluted than you mobile phase. If your target compound is less retarded than the solvent you dissolved the sample it can cause a change in the refractive index in the flow cell. That can cause a positive or negative Peak. But we need more information.
Gerhard Kratz, Kratz_Gerhard@web.de
Another common reason for a negative peak is the use of the 'background subtract' wavelength (most commonly set up on Agilent / HP systems and called the reference wavelength). If the eluting species absorbs strongly at that wavelength and relatively poorly at your recorded wavelength you will see a negative peak. This one is easy to check - just re-run the sample with the reference turned off.
4 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry