Instrument conditions:
Splitless mode
Injector: 280'C
Detector: 280'C
Injection volume: 1.0uL
Column Flow: 1.0mL/min
Total Flow: 23.3mL/min
Starting pressure: 7.7psi
Purge Flow: 20.0mL/min @0.0 min
Source:280'C
Quad:180'C
Instrument was running fine (m/z 69~500,000 in autotune, 1,4-DCB-d4~180,000 IS response, no leaks) until source maintenance. After source cleaning the autotune and DFTPP tune looked fine (same abundance for m/z 69 and no leaks) but my 1,4-DCB-d4 IS response has decreased 10 fold (lucky to break 20,000). I can get more sensitivity out of the instrument by increasing the purge time to 0.75 minutes but the chromatography for the early eluters suffers greatly (mostly blobs rather than sharp peaks). I've also noticed that the ramping for my repeller, entrance lens and entrance lens offset look a little funny (not exactly Gaussian shaped and reppeler is maxing out) Could this be a source issue or something with the injector? I've done injector maintenance twice to make sure there were no loose connections, but I did just start using a bore brush to clean the inside of the port. Possible active sites? Thanks for your time.