nimbi63 » Wed Jul 24, 2013 4:34 pm
With both instruments closed, run IQT (Installation Qualification Tests) to see if any programs have been changed or moved.
Open the Chem32\2\Methods folder and copy a calibrated method to another folder. Open Instrument 1 and select Load Method, browse to this folder and load the method. Does the cal table appear? If so, the method is fine.
Browse to Chem32\2 with instrument 2 closed, and rename the Config.Reg file to Config.Bak (if a config.bak exists, rename it to Config.Old first). Start ChemStation and load a method, does the cal table exist? If so, on the command line enter: RestoreColumns <This command will copy any columns you have entered from Config.bak to the system's config register, and create a new config.reg the next time ChemStation closes.
If this fails, close ChemStation, browse to Chem32\2\ and rename the Temp directory to Temp.Old. Create a new, empty Temp directory in Chem32\2, then restart ChemStation.