Name / B.P./ Spec Limit
IPA /82/ 2%
MeCN /82/ 2%
CPM-Cl /89/ 20 ppm
CPM-I /142/ 20 ppm
The problems we have:
1. We tried several columns and diluents (DMF, DMSO, Tolune, etc). CPM-I is always co-elute with diluent.
2. The FID sensitivity for CPM-Cl/CPM-I seems very low. We also found that the CPM-I being such a high boiler did not have consistent area counts when injected at low levels.
3. In order to be able to detect low level of CPM-Cl/CPM-I, sample amount needs to be very high (e.g. 500 mg/mL). However, the high amount of MeCN in the sample caused severe carry over in the headspace requiring service calls to change out the 6 port valve and sample loop in order to get it out.
4. We tried to use two HS-GC methods, one for high level of IPA/MeCN (pretty straightforward), another one for low level CPM-Cl/CPM-I, but we still have the same co-elute issue for CPM-I with diluent DMF, DMSO as said above and same carryover issue of MeCN when using MeCN as diluent.
5. We also tried direct injection GC for CPM-Cl/CPM-I, the LOD is ~8 ug/mL with peak area and height only 5 PA. That is 800 ppm if the sample conc is 10 mg/mL (By the way, typically how much of a sample (ex. 1 mg/mL) can be injected directly onto the GC? do not care the peak distortion of API).
Any suggestions would be greatly appreciated!