negative baseline
Posted: Tue Jul 09, 2013 1:06 pm
by Shiveirou Raomai
I have been trying to run hplc-UV for quite a long time. but everytime baseline become negative.what could be the reason?can anyone please suggest some solutions?
wavelength 203nm,flow rate 1ml/min, solvent-water and acetonitrile at different ratio. initially there was no problem but few weeks back negative baseline appeared even while monitoring and running blank. I have changed the column twice but to no avail.
Re: negative baseline
Posted: Tue Jul 09, 2013 2:31 pm
by bisnettrj2
Post your analysis parameters, your instrument type, your UV detection wavelength, your mobile phase, flow rate, etc.
Posted: Wed Jul 10, 2013 5:48 am
by Gerhard Kratz
Especiallyat low wavelength you feel that your system is not in an equillibrium. Maybe it was not flushed well and you have still a solvent in the system what does not belong to your mobile phase, or you have a temperature problem, or your gradient pump system is not working propper. But without details it is impossible to give you better support.