I have been trying to run hplc-UV for quite a long time. but everytime baseline become negative.what could be the reason?can anyone please suggest some solutions?
wavelength 203nm,flow rate 1ml/min, solvent-water and acetonitrile at different ratio. initially there was no problem but few weeks back negative baseline appeared even while monitoring and running blank. I have changed the column twice but to no avail.
Post your analysis parameters, your instrument type, your UV detection wavelength, your mobile phase, flow rate, etc.
Time flies like an arrow. Fruit flies like a banana.
Especiallyat low wavelength you feel that your system is not in an equillibrium. Maybe it was not flushed well and you have still a solvent in the system what does not belong to your mobile phase, or you have a temperature problem, or your gradient pump system is not working propper. But without details it is impossible to give you better support.
Gerhard Kratz, Kratz_Gerhard@web.de
3 posts Page 1 of 1

Who is online

In total there are 361 users online :: 1 registered, 0 hidden and 360 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 360 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry