Perkin Elmer Autosystem communication error/ codes
Posted: Wed Jul 03, 2013 3:14 pm
by HannuN
Hi.
We have problems with Perkin Elmer Autosystem XL (Arnel) GC havin Turbochrom control and integration program and "Black box " connection between PC and GC.
GC is telling us: DTF
And turbochrom continues: Seize error (993) or seize error 35 at A.
My guestions: I. where can I found explanation to these error codes?
Question II: what might be the problem and is there any easy (read cheap) solution to that.
Yours.
Hannu Nykänen
Univ. of Jyväskylä, Finland
Re: Perkin Elmer Autosystem communication error/ codes
Posted: Wed Jul 03, 2013 6:08 pm
by rb6banjo
I'm not sure where you can find a master list of the codes but I too have an Arnel. DTF means the detector support gases (one or more) are not what they should be.
I had to change a PPC module a couple of weeks ago because I couldn't get a DTF error message to clear. It was the air flow controller on my second FID. Swapped it out with another and it has been working beautifully since.
Posted: Mon Jul 15, 2013 9:57 pm
by cbwhipp
DTF is the detector flow. My guess would be it is your Air for the FID? Make sure you have between 80-90 psi going to the GC. If your gas supplies are good, than it is likely that the PPC module is bad and will need to be replaced.