Hey I was wondering if anyone knows how to change the default units when using empower.

So i am running a residual solvent method and one of my peaks looks great has >100 data points and yet the area is around 1 pA*sec (?) is there a way to convert this in some default or would I only be able to make a custom calculation and multiply it by a lot?
Y-units are pA, areas displayed SAY µV*sec.
It really shouldn't matter so long as your calibrators and samples use the same units. When you divide by the response factor to get concentration (or mass), the area units cancel out.
-- Tom Jupille